Aug 16, 2025 at 3:55 PM#1

I have a certificate reporting 99.2% purity and 6.1mg content on a vial sold as 5mg, and I cannot work out whether those two numbers are consistent or contradictory.

What I actually want to know is whether a purity figure and a content figure can disagree without one of them being wrong.

Happy to be told the question itself is wrong.

42 12BiostatsBrad, PeptideSynthNJ, Dr.KarenChen and 39 others
Reply Quote Save Share Report
PeptideChemSF
Senior Member
1,890
9,012
Jan 2024
San Francisco, CA
Aug 16, 2025 at 5:18 PM#2

Short answer first, then the reasoning. Inter-laboratory variation comes from real methodological differences, not sloppiness: which reference standard was used and how it was itself calibrated, the gradient and column, the detection wavelength, and where the integration baseline was drawn. A percentage point or two between competent labs is expected. What is not expected is a large gap, and the usual explanation there is that one report is not of the batch in your hand.

41 11dave_SLC, FDA_TrackerJim, ricardo_MIA and 38 others
Reply Quote Save Share Report
RegAffairsDC
Member
678
3,456
May 2024
Washington, DC
Aug 16, 2025 at 6:40 PM#3
PeptideChemSF said:
Inter-laboratory variation comes from real methodological differences, not sloppiness: which reference standard was used and how it was itself…

Agreed, and mass spec answers a question purity does not: identity. High purity tells you one thing dominates the chromatogram, not that it is the molecule you ordered. A mass within about a dalton of expected is the identity check.

Last edited: Aug 16, 2025 at 8:40 PM
40 10hannah_MT, Dr.SportsMedIN, amy_econ_NJ and 37 others
Reply Quote Save Share Report

Sigma-Aldrich — Research-Grade Standards

Certified reference materials, analytical reagents, and research-grade standards for peptide verification. Trusted by laboratories worldwide.

Shop Reference Standards
quinn_sf
Member
489
2,123
Jun 2024
San Francisco, CA
Aug 16, 2025 at 8:03 PM#4
DebRD_ATL said:
I have a certificate reporting 99.2% purity and 6.1mg content on a vial sold as 5mg, and I cannot work out whether those two numbers are consistent or…

Same position here, arrived at the long way round. They are not contradictory, they measure different things, and this is the single most common misreading of a certificate. Purity is a relative area measurement: the main peak as a percentage of total peak area in the chromatogram. Content is an absolute mass measurement against a reference standard. A vial can be 99% pure and contain 6.1mg — pure material, overfilled — or 99% pure and contain 3mg, which is pure material and a short fill. Purity tells you what fraction of what is there is the right molecule; content tells you how much of it there is. You need both, and a certificate with only purity is telling you half the story.

39 9andrew_nyc, Dr.EndoEP, GraceAZ_72 and 36 others
Reply Quote Save Share Report
SarahChen_PharmD
VIP Member
4,567
22,341
Dec 2023
San Diego, CA
Aug 17, 2025 at 4:13 AM#5

From the other side of the consultation, briefly.

Reconstituted peptide stability data for certificates of analysis: I had Janoshik test my reconstituted vial at day 1, day 14, and day 28 stored in the refrigerator.

Day 1: 100.1% purity
Day 14: 99.6% purity
Day 28: 97.3% purity

Minimal degradation over 28 days when properly refrigerated. This confirms the standard 28-day beyond-use date is appropriate. Beyond 28 days, I'd reconstitute a fresh vial to be safe.

Last edited: Aug 17, 2025 at 7:13 AM
38 8robert_kc, dan_philly, MeganSA_TX and 35 others
Reply Quote Save Share Report
1239

Similar Threads

HPLC vs UPLC for peptide purity — method comparison study18 replies
Mass spectrometry for peptide identity verification — ESI-MS guide7 replies
Endotoxin testing methods — LAL vs recombinant Factor C17 replies
Red flags on COAs — how to spot a fake certificate5 replies
USP reference standards for peptide verification — sourcing guide3 replies
ForumsNewTrendingMembersAccount

Log In

Forgot password?
No account? Register